Hi, It's been a while since I used the package myself, so I am not sure what will happen if there isn't enough information to fully constrain the calibration. I think it will lead to high uncertainty, weird drifts in temperature, etc. I don't think the package will raise an error.
If you have only a single bath it will be basically impossible to calibrate the data properly. To get a ballpark estimation you can calibrate In a double-ended setup you can probably get away with having a single bath, but two would give you more certainty in the accuracy of your setup. You could also make use of opportunistic "baths". I.e. a section of the fiber in a medium with a relatively constant and homogeneous temperature in which you place the fiber, along with a temperature sensor with a separate logger. It will depend on your deployment what this can be. Internal reference sections of the interrogator are basically useless for calibration purposes. The only use for this as end-user of the interrogator is to determine the relationship between signal strength and measurement noise of your device.
If you have the two fibers co-located in some way, you can correct for this by assuming that these sections of fiber are at the same temperature, giving you extra constraints. Could you tell us a bit more about who you are and what you're using DTS for? It's nice to have an idea who we're talking to! Lastly, I would be up for providing commercial services such as consulting on the use of this software & calibrating DTS data and/or maintaining this software if you are interested (and I assume @bdestombe would be as well). |
Beta Was this translation helpful? Give feedback.
Uh oh!
There was an error while loading. Please reload this page.
-
Hi!
Great package idea, I'm just going to get started using it.
How does the code behave if there is not sufficient information available to fully solve for all calibration coefficients?
e.e. 3 independent reference measurements of temperature are not available
There are two key examples for my use cases:
where it is not possible to use sufficient calibration baths (perhaps one internal to the interrogator, one outside the interrogator, and that is it - the far end of the fibre is inaccessible).
where the fibre experiences transient attenuations at point locations (bad splices, damage etc.) but it is not possible to have additional calibration baths either side of the loss.
The temperature calculated on the device can show discrepancy in some of these scenarios. But would this package provide a different answer when solving with insufficient information? And why?
Thanks in advance.
Beta Was this translation helpful? Give feedback.